Determining the Equilibrium Constant of a Chemical Reaction
Purpose To determine the equilibrium constant, Kc (also called Kf) for the formation of a complex ion by measuring equilibrium concentrations of the reacting species involved. Introduction Many methods may be used to determine the equilibrium constant for a given system. A pH meter may be used to determine the acid dissociation constant, Ka, and it is also possible to determine the solubility product equilibrium constant, Ksp. However in this experiment the equilibrium constant, Kc, for the formation of a complex ion will be determined by measuring the concentrations of reactants in the reaction below. The equilibrium expression is also shown below.
Fe3+(aq) + SCN- (aq) FeSCN2+ (aq) πΎπ = [πΉπππΆπ2+]
[πΉπ3+][ππΆπβ]
The FeSCN2+ complex ion is a strongly colored species; therefore the reaction can be investigated using spectroscopy. In part A, a calibration curve will be constructed using FeSCN2+ solutions of known concentration. Each standard solution contains a large excess of Fe3+ to ensure all SCN- in solution forms the FeSCN2+ complex. Therefore, the concentration of FeSCN2+ at equilibrium is equal to the initial concentration of SCN-. The calibration curve will be used to determine the equilibrium concentration of FeSCN2+ of each sample (prepared in part B) containing similar concentrations of Fe3+ and SCN- at equilibrium. In part B, mixtures containing similar concentrations of Fe3+ and SCN- will be prepared. Each solution will be allowed to reach equilibrium such that the following mass balance equations (1) and (2) are valid. With these equations and the initial concentration of the two ions the equilibrium concentrations can be determined.
[Fe3+]initial = [Fe3+]eq + [FeSCN2+]eq (1)
[SCN- ] initial = [SCN- ] eq + [FeSCN2+] eq (2)
An alternate and probably more familiar way to determine equilibrium concentrations is the ICE box method. The ICE box method uses the same equations as above but arranges them in a chart like format.
Fe3+(aq) + SCN- (aq) FeSCN2+ (aq)
Initial I
Change ο
Equilibrium Eq
The initial concentrations of iron Fe3+ and thiocyanate SCN- ions can be easily determined from the concentration of the stock solutions. The absorbance of the iron thiocyanate ion will be determined by spectroscopy and compared with the calibration curve prepared in part A to determine the equilibrium concentration of the ion. The equilibrium concentrations can be found using the ICE box chart.
Short Review of Beerβs Law & Spectroscopy Beerβs Law is defined below in equation: A = a b c In the equation, A is the measured absorbance of the sample, b is the path length of light through the sample, c is the concentration of the sample, and a is a constant that depends on both wavelength and substance. A linear calibration curve, or Beerβs Law plot, of absorbance (y-axis) versus concentration (x-axis) can be constructed from the absorbance and concentration data of standard solutions. A prepared calibration curve can be used to determine the concentration of an unknown solution from its measured absorbance. Experimental Procedure: General Instructions β Read this BEFORE proceeding to Parts A and B. 1. The SpectroVis Spectrometers and LabQuest units will be used for this experiment. The
spectrometers need to be calibrated using the blank solutions for Part A and Part B respectively. See the procedure below for instructions on calibrating the spectrometer. The wavelength to be monitored for all samples is 460 nm.
2. There are four solutions used in this experiment. Do not take more solution than you need and do
not cross contaminate the solutions.
ο· Obtain two burets to deliver the 0.004 M Fe3+ and 0.004 M SCN- solutions. The burets must be clean prior to use. Be sure to rinse the buret with a few mL of the solution prior using.
ο· Use your 10.00 mL pipet for the 0.1 M Fe3+. ο· Use a graduated cylinder for the 1.0 M HNO3 and DI water. You can use a plastic pipet to
accurately fill the cylinder.
For each solution record the exact volumes you deliver in your data tables (and notebook) with the correct number of significant digits.
3. Prepare all solutions in test tubes large enough to hold 20 mL (provided). 4. Thoroughly mix all solutions by stoppering the tube and inverting at least five times. If the
solutions are not properly mixed there will be significant error in your results. 5. DO NOT PREPARE solutions for part B until you have completed Part A. Instrument Setup and Calibration 1. In the menu bar select Sensors then choose Data Collection. Select the dropdown arrow next to
the Mode box and choose the Time Based option. Leave other settings as is. Tap OK. 2. In the menu bar select Sensors, then choose Calibrate, and select USB:Spectrometer. A
calibration page should appear. Tap OK. 3. After the warm-up period obtain a cuvette filled with water to calibrate the spectrometer. Tap
the Finish Calibration button. When the unit states calibration complete tap OK. 4. Back on the main screen, tap the panel labeled USB: Abs. Select Change Wavelength from the
menu. 5. For the Selected Wavelength enter 460 nm. Tap OK when done.
Part A: Construction of Beerβs Law Plot
1. Prepare blank A and samples 1A β 4A using the 0.10 M Fe3+ solution according to the volumes listed below. Since the measured volumes may vary slightly from the volumes in the table below, record the exact volume of each reagent used in Table 1 on page 5.
Sample Volume (mL) of
0.10 M Fe3+ Volume (mL) of
0.004 M SCN- Volume (mL) of
1.0 M HNO3 Volume (mL) of Deionized H2O
Blank A 10.00 0.00 5.0 5.0
1A 10.00 0.30 5.0 4.7
2A 10.00 0.60 5.0 4.4
3A 10.00 1.20 5.0 3.8
4A 10.00 2.00 5.0 3.0
2. Calibrate the spectrometer with water. When calibrated obtain an absorbance reading for the
Blank A solution. This will be used to correct for any background absorbance in the samples. 3. Record the absorbance of each sample in Table 2 (in your notebook). 4. Calculate the concentration of [FeSCN2+]eq and place the value in Table 2 (in your notebook). NOTE: a) [FeSCN2+]eq = [SCN- ]initial. This is true only for part A